BACKGROUND: The accuracy of oscillometric blood pressure (BP) measuring devices determined in resting conditions may be distorted in those under hemodialysis (HD). METHODS: The evaluation of accuracy of the Omron HEM-907 was performed in accordance with the International Protocol for validation of BP measuring devices (IP). Out of a group of 49 patients with end-stage renal disease (ESRD) (56.7 +/- 15.6 years), four groups of 33 patients each were selected to ascertain their BP before and after HD. RESULTS: The Omron HEM-907 fulfilled the IP accuracy requirements for systolic BP measurements before and after HD. In the case of diastolic BP before HD, the Omron HEM-907 did not fulfill both requirements for phase 2.2 IP. In only 21 out of 33 patients, the difference between the mean of two observers and the Omron HEM-907 device (Diff) was not more than 5 mmHg, while the requirement was 22 patients. In seven patients, all three Diff were over 5 mmHg, while the accepted limit is!